Canonical Question

Resp – Oximetry

V5 C9.i Historical V4 F12.i 1 appearance

Master answer

Pulse oximeters assumes that only oxy-Hb and deoxy-Hb are present in the blood. It cannot differentiate other species of haemoglobin such as carboxyhaemoglobin (carboxy-Hb) or methaemoglobin (met-Hb) because it only uses two wavelengths of light (660 nm and 940 nm).

Co-oximeter:

Image Source: Maths, Physics and Clinical Measurement for Anaesthesia and Intensive Care. Cambridge University Press.

Mechanism / Measurement

\[ FO_2Hb \; = \; {{oxy-Hb} \over {oxy-Hb \; + \; deoxy-Hb \; + \; carboxy-Hb \; + \; met-Hb}} \]

Advantages:

Limitations of co-oximetry:

Limitations of pulse oximetry:

PATIENT FACTORS
Low or High SpO2Low SpO2Normal or High SpO2
Met-Hb
Sulph-Hb
Poor perfusion of finger
Movement artifact
Venous pulsations
Fingernail polish
Intravenous pigmented dyes
Haemoglobinopathy
Anaemia with co-existing hypoxia
Carbon Monoxide poisoning
EQUIPMENT FACTORS
– Ambient light interference
– Poorly fitting probe
– Assay calibrated using healthy volunteers only down to SpO2 80%. Unknown significance if tested SpO2 less than 80%
PHYSIOLOGICAL FACTORS
– Due to O2 dissociation curve, insensitive to changes above PaO2 80mmHg
– Does not measure tissue oxygenation

JC 2020

Exam appearances

ExamExact wordingRelationshipSuccess
2018A Q08 Outline the principle of co-oximetry (40% of marks), describe what a co-oximeter is able to measure (30% of marks), and compare its limitations to those of a pulse oximeter (30% of marks). historical_member