Resp – Capnography
Variant family: at least one exam appearance has different scope. The appearance record controls whether this master answer is safe to display on that Past Paper.
Master answer
Principles
- Beer Lambert law: At a given wavelength, the amount of infrared radiation absorbed by gas is proportional to the concentration of gas present
- CO2 is a heteronucleic molecule, and so absorbs infra-red light
- Infrared light shone across a sample of gas
- Narrow band light emitted from infrared source
- Band frequency chosen which fits the peak absorption frequency of CO2
- 4.23micrometres
- Shone across gas, and absorbed at a detector
- Detector emits signal to analyser -> screen
- CO2 reading outputted is inversely proportional to CO2 present in sample, as per Beer-Lambert Law
- Narrow band light emitted from infrared source
- Gas can be sampled at the patient within the ventilation circuit (in-line), or in a sample of gas diverted away to a separate analysing chamber (sidestream)
Sources of error and limitations
- Sampling
- Entrainment of atmospheric gas if leak in sidestream line
- Occlusion of sidestream line causes loss of gas sampling
- Water condensation absorbs IR light -> erroneously high ETCO2
- Modern capnography includes a water trap and heater (to reduce condensation)
- Calibration
- Incorrect calibration of analyser
- Interference
- Other gases (notably N2O) have a similar absorption spectrum
- Presence may falsely elevate measured ETCO2 (esp. if infrared frequency band too broad)
- Presence of other gases causes ‘collision broadening’
- Absorption spectrum of CO2 is broadened
- Pressure
- Partial pressure, rather than percentage composition, is measured
- If pressure ↓ (e.g. by suction drawing gas into sampling chamber), erroneously low measured ETCO2
- Sampling chamber
- If too large, mixing of gas between respiratory cycles -> compression of waveform and erroneously low measured ETCO2
Mooney 2016
Exam appearances
| Exam | Exact wording | Relationship | Success |
|---|---|---|---|
| 2019B Q10 | Describe the principles of capnography, including calibration, sources of error and limitations. | historical_member | — |
| 2010A Q04 | Describe the underlying principles involved in the measurement of end tidal CO2 (by infrared analysis), including sources of error and interference. | historical_member | — |
| 2015B Q09 | Describe the principles of measurement of end-tidal CO2, including the sources of error. | historical_member | — |
| 2023A Q03 | Outline the principles of measurement of end-tidal CO2 using infrared radiation (25% of Marks). Describe the potential sources of error when using this modality and how they may be mitigated (75% of Marks). | historical_member | — |
| 2026A Q11 | a) Outline the scientific principles that apply to the measurement of end-tidal carbon dioxide using capnography. Include the techniques of sampling in your answer (30% of marks). b) Describe a normal capnograph waveform and its features (20% of marks). A diagram may assist you with your answer. c) Outline the ventilation and perfusion information that can be derived from the capnograph waveform (50% of marks). | variant | 48.20% |